Example of the Bias Report and Linearity Study

Use the Variability Chart platform to evaluate the bias and linearity of a new measurement system. Five parts were chosen throughout the operating range of the measurement system, based on documented process variation. Each part was measured by layout inspection to determine its reference value. Each part was then measured twelve times by the lead operator. The parts were selected at random during the day.

Bias Report

Use the Bias Report option to examine the overall and individual measurement bias.

1. Select Help > Sample Data Folder and open Variability Data/MSALinearity.jmp.

2. Select Analyze > Quality and Process > Variability / Attribute Gauge Chart.

3. For Chart Type, select Variability.

4. Check that the value for the Sigma Multiplier is 6.

5. Select Response and click Y, Response.

6. Select Standard and click Standard.

7. Select Part and click X, Grouping.

8. Click OK.

9. Click the Variability Gauge Analysis for Response red triangle and select Gauge Studies > Bias Report.

Figure 6.11 Measurement Bias Report 

Measurement Bias Report

The bias (Response minus Standard) is calculated for every measurement. The Overall Measurement Bias Report shows a histogram of the bias and a t test to see whether the average bias is equal to 0. You can see that the Average Bias is not zero, it is -0.0533. However, zero is contained within the confidence interval (-0.1152,0.0085), which means that the Average Bias is not significantly different from 0. Using a significance level of 0.05, you can see that the p-value is greater than 0.05, which also shows that the Average Bias is not significantly different from 0.

The Measurement Bias Report by Standard shows average bias values for each part. The bias averages are plotted on the graph along with the actual bias values for every part, so that you can see the spread. In this example, part number 1 (with a standard value of 2) is biased high and parts 4 and 5 (with standard values of 8 and 10) are biased low.

Tip: To see confidence intervals for the bias, right-click in the table and select Columns > Lower 95% and Columns > Upper 95%.

Linearity Study

Use the Linearity Study option to determine whether there is a significant relationship between the size of the parts and the operator’s ability to measure them. First, you must specify a historical process sigma.

1. Click the Variability Gauge Analysis for Response red triangle and select Gauge Studies > Edit MSA Metadata.

2. In the Edit MSA Metadata window, enter 2.488105 as the value for the Historical Process Sigma. This is the standard deviation of the response. The Historical Process Sigma is used to calculate the Process Variation.

3. Click OK.

4. Click the Variability Gauge Analysis for Response red triangle and select Gauge Studies > Linearity Study.

Figure 6.12 Linearity Study 

Linearity Study

Note the following:

• The slope is -0.131667. This value appears as part of the equation below the graph, and also in the third table.

• The p-value associated with the test on the slope is quite small, <.0001. The t test for the slope is testing whether the bias changes with the standard value.

Because the p-value is small, you can conclude that there is a significant linear relationship between the size of the parts and the operator’s ability to measure them. You can also see this in the graph. If the part or standard value is small, the bias is high, and vice versa.