DRAFT help
Publication date: 01/13/2026
Statistical Details for the Gauge Attribute Chart
In the Gauge Attribute chart that plots all X, Grouping variables on the horizontal axis, the percent Agreement is calculated as follows:
Equation shown here
In the Gauge Attribute chart that plots all Y, Response variables on the horizontal axis, the percent Agreement is calculated as follows:
Equation shown here
where
n = number of parts (grouping variables)
ri = number of reps for part i (i = 1,...,n)
m = number of raters
k = number of levels
Ni = m × ri. Number of ratings on part i (i = 1,...,n). This includes responses for all raters, and repeat ratings on a part. For example, if part i is measured 3 times by each of 3 raters, then Ni is 3 × 3 = 9.
For example, consider the following table of data for three raters, each having three replicates for one part.
Table 7.1 Three Replicates for Raters A, B, and C
A
B
C
1
1
1
1
2
1
1
0
3
0
0
0
Using this table, you can make these calculations:
% Agreement =Equation shown here
% Agreement [rater A] = percent Agreement [rater B] = Equation shown here and
% Agreement [rater C] =Equation shown here
